We all are used to seeing a pair of high voltage power supply caps wired in series with ballast resistors across each, usually on the first filter stage of the amplifier. It's normally done to achieve sufficiently high votage rating for cap values chosen, where a single cap doesn't have high enough voltage rating. I normally have a look at those during service work, to see that half the supply voltage is across each cap in series, as it should be. Usually they are.
In the present instance, where I'm servicing a vintage Hewlett-Packard 712C High Voltage Power supply, perfect for tube amp work (0-500VDC @ 0-200mA, 0 to -150VDC Bias, 6.3VAC CT @ 15A heater supply), with date codes from 1971 (unit came from Sylvania). I'm inside servicing the Constant Current Comparator circuit, and recapped most of the aged electrolytics. The Output caps following the series voltage regulator used a pair of 80uF/300V caps, twist-tab style, and I fitted in a pair of 100uF/400V parts, with a minor lead spacing adjustment & O-Rings glued in for mechanical stability.
After finding a small current surge when I'd switch on the HV supply, the output would overshoot (to 500V from a 300V setting!), then settle down to desired setting after a few seconds, completely uncharacteristic of lab supplies! After checking a few things, I measured across the 100k ballast resistors of the two 100uF caps and found a huge imbalance (225VDC & 75VDC for a 300VDC setting), under no load on the output. Pulled the caps, checked them on one of the two bridges I have here, reading 92uF & 90uF, same DF reading. Close enough match, but whose capacitive reactance difference shouldn't be enough to cause that much imbalance. Brand new parts.
I looked at the original parts I had removed. One cap was almost half the value of the other @ 120Hz, and at 1kHz, 47nF instead of around 80uF. So, no sense in putting those caps back into place to see that order is restored. I won't go into my pondering on why, after installing a pair of 47uF caps temporarily has now yielded no output and spun me off into assessing the SCR control circuitry on the front end, and the series regulator and other control loops feeding in the system. My new headache.
I'm wondering what insures proper voltage balance. Now, I did state NO LOAD on the output of the supply, while observing this gross imbalance. Is that the key to having proper balance on the front end stacked buss caps? In that instance, you DO have charge current flowing thru them. Not so in my case where they're merely smoothing filters (no load), and would have a little bit under DC load. Makes me go hmmmmmm..........
In the present instance, where I'm servicing a vintage Hewlett-Packard 712C High Voltage Power supply, perfect for tube amp work (0-500VDC @ 0-200mA, 0 to -150VDC Bias, 6.3VAC CT @ 15A heater supply), with date codes from 1971 (unit came from Sylvania). I'm inside servicing the Constant Current Comparator circuit, and recapped most of the aged electrolytics. The Output caps following the series voltage regulator used a pair of 80uF/300V caps, twist-tab style, and I fitted in a pair of 100uF/400V parts, with a minor lead spacing adjustment & O-Rings glued in for mechanical stability.
After finding a small current surge when I'd switch on the HV supply, the output would overshoot (to 500V from a 300V setting!), then settle down to desired setting after a few seconds, completely uncharacteristic of lab supplies! After checking a few things, I measured across the 100k ballast resistors of the two 100uF caps and found a huge imbalance (225VDC & 75VDC for a 300VDC setting), under no load on the output. Pulled the caps, checked them on one of the two bridges I have here, reading 92uF & 90uF, same DF reading. Close enough match, but whose capacitive reactance difference shouldn't be enough to cause that much imbalance. Brand new parts.
I looked at the original parts I had removed. One cap was almost half the value of the other @ 120Hz, and at 1kHz, 47nF instead of around 80uF. So, no sense in putting those caps back into place to see that order is restored. I won't go into my pondering on why, after installing a pair of 47uF caps temporarily has now yielded no output and spun me off into assessing the SCR control circuitry on the front end, and the series regulator and other control loops feeding in the system. My new headache.
I'm wondering what insures proper voltage balance. Now, I did state NO LOAD on the output of the supply, while observing this gross imbalance. Is that the key to having proper balance on the front end stacked buss caps? In that instance, you DO have charge current flowing thru them. Not so in my case where they're merely smoothing filters (no load), and would have a little bit under DC load. Makes me go hmmmmmm..........