A buddy of mine at work has a BK Precision 886 LCR meter, I recently discovered, so I borrowed it and did a correlation with my Extech 380193.
The BK is a higher end (at least price-wise) model. It does measure more frequencies (100, 120, 1000, 10000, 100000 Hz) and also provides impedance and phase angle measurements.
Most measurements correlate to within a fraction of a percent across a wide range of values.
I hope this also serves to clarify some of the measurements that the Extech outputs, since the documentation on that meter is so thin.
Based on this study, I conclude that the "R" in the primary (L,C,R) functions of the Extech is the equivalent series resistance (ESR). The secondary R that the Extech provides is ESR and EPR (equivalent parallel resistance) in the series and parallel modes, respectively. The difference between the primary ESR and secondary ESR is unclear. The measurements correlate well, although the primary value is consistently lower by a hundred or so Ohms at 1000 Hz (at 120 Hz, they are much closer and the primary value may be higher or lower). As far as I can tell, the primary mode seems to be done with an internal calibration, as the measurement tends to start at the higher value (close to the secondary value) and settle in at the lower value after it chews on it for a few seconds. The primary mode also tends to continuously recalculate, whereas the secondary mode does not. I assume this means that the primary mode is more accurate.
Also, the ESR at 120 Hz correlates well with the DCR measured by the BK, so for all practical purposes, the R at 120 Hz measured by the Extech can be treated as the DCR for pickups. Some of the correlations are shown below. The grey dotted line represents y=x. The correlation for the ESR is the least robust, but still not bad. It should be noted that the Extech actually seems to handle ESR measurement better than the BK. The BK in fact cannot handle pickup ESRs greater than about 9500 Ohms. That's why that comparison is truncated. The Extech seems to handle ESR (assuming that is what this measurement actually is) across the board.
I measured a representative range of my models spanning a wide range of L and R and with varying degrees of eddy current loading.
I hope you find this useful. I'm certainly very comfortable with the Extech numbers at this point. The inductance values are spot on at both 120 and 1000 Hz.
The BK is a higher end (at least price-wise) model. It does measure more frequencies (100, 120, 1000, 10000, 100000 Hz) and also provides impedance and phase angle measurements.
Most measurements correlate to within a fraction of a percent across a wide range of values.
I hope this also serves to clarify some of the measurements that the Extech outputs, since the documentation on that meter is so thin.
Based on this study, I conclude that the "R" in the primary (L,C,R) functions of the Extech is the equivalent series resistance (ESR). The secondary R that the Extech provides is ESR and EPR (equivalent parallel resistance) in the series and parallel modes, respectively. The difference between the primary ESR and secondary ESR is unclear. The measurements correlate well, although the primary value is consistently lower by a hundred or so Ohms at 1000 Hz (at 120 Hz, they are much closer and the primary value may be higher or lower). As far as I can tell, the primary mode seems to be done with an internal calibration, as the measurement tends to start at the higher value (close to the secondary value) and settle in at the lower value after it chews on it for a few seconds. The primary mode also tends to continuously recalculate, whereas the secondary mode does not. I assume this means that the primary mode is more accurate.
Also, the ESR at 120 Hz correlates well with the DCR measured by the BK, so for all practical purposes, the R at 120 Hz measured by the Extech can be treated as the DCR for pickups. Some of the correlations are shown below. The grey dotted line represents y=x. The correlation for the ESR is the least robust, but still not bad. It should be noted that the Extech actually seems to handle ESR measurement better than the BK. The BK in fact cannot handle pickup ESRs greater than about 9500 Ohms. That's why that comparison is truncated. The Extech seems to handle ESR (assuming that is what this measurement actually is) across the board.
I measured a representative range of my models spanning a wide range of L and R and with varying degrees of eddy current loading.
I hope you find this useful. I'm certainly very comfortable with the Extech numbers at this point. The inductance values are spot on at both 120 and 1000 Hz.