The reason I said "the sockets" were the culprit, is that I disconnected all the wiring from the sockets to the rest of the circuit. So in the schem, I disconnected the wires between the plate switch (S1) and the sockets, as well as the filament wiring (black & brn/black) from S3 and S6 and T2. Totally isolated the socket panel.
In that state, with the sockets and their wiring just hanging there, not connected to anything, the 8050A meter still registered the same leakage among the sockets. No leakage detected among the switch contacts. Powering the unit up with sockets disconnected like this, the tester itself detected ZERO leakage in any switch position. So the phenolic switches don't seem to be the problem, it's either the sockets or socket wiring or both.
I can't do more testing on this right now, because of course now the humidity is around 45-50%, so the whole thing is working just fine.
In that state, with the sockets and their wiring just hanging there, not connected to anything, the 8050A meter still registered the same leakage among the sockets. No leakage detected among the switch contacts. Powering the unit up with sockets disconnected like this, the tester itself detected ZERO leakage in any switch position. So the phenolic switches don't seem to be the problem, it's either the sockets or socket wiring or both.
I can't do more testing on this right now, because of course now the humidity is around 45-50%, so the whole thing is working just fine.