So you can get a sweep, but it goes from upper left to lower right? Maybe the CRT has shifted. Some scopes have a trace rotation adjustment, but on this one maybe you just need to loosen the CRT clamp and physically rotate the CRT so the sweep is horizontal.
Ad Widget
Collapse
Announcement
Collapse
No announcement yet.
Heathkit ol-1 oscilloscope help. Can't get solid spot during setup procedure.
Collapse
X
-
Originally posted by JoeM View PostSo you can get a sweep, but it goes from upper left to lower right? Maybe the CRT has shifted. Some scopes have a trace rotation adjustment, but on this one maybe you just need to loosen the CRT clamp and physically rotate the CRT so the sweep is horizontal.
Comment
-
Find the power supply caps, ideally use another scope to look at what they are doing. Otherwise, measure the DC voltage ion them and see if it seems reasonqable, but mainly measure them for AC voltage to see how much rippple there is. Caps are suspects, at 50 years old certainly the electrolytics, but also the film caps. When you recapped it, did you replace ALL the caps, or just e-caps?
have you tried different tubes?
Wait, you say you have all the documentation? Like including the schematic? Geez scan that and post it.Education is what you're left with after you have forgotten what you have learned.
Comment
-
I only replaced the electrolytics with fresh caps and brought the unit up slowly on the variac. Haven't touched the others. I will look at the power supply. If thats not a problem, once I have replaced all out of value resistors if there is still an issue I will have to look at the other caps. Hate to touch those cool old yellow spragues. Anyone restoring tweed or brown Fenders would go nuts for em!
I don't have the info on this computer but amazingly I found a link while scanning a forum somewhere with the full manual and schematic. Can't find it in a Google search. I will post them tomorrow. My workshop and studio are at my Dad's house just a couple miles away and that is where the gear and files are. As always thanks for your time. Hasta manana!
Comment
-
I don't necessarily want you to replace them without question, but if they are leaking, that makes DC go where it is not wanted. And at 50 years, that is always a concern.
Drifted resistors are more likely to alter performance, rather than prevent it. The voltage scales might be off, or the linearity funny so the waveform distorts, or the sweep rates cockeyed, but unless the resistors are WAAAAAY off, if everything else is OK, then the thing should function. Just like a guitar amp.
The deflection is a pretty simple push-pull affair, but a dead tube stage can easily kill the signal.
And while this is a visual device, if you feed some audio signal into it, you can still use a signal tracer to see if it is passing through the scope stages.
If you google "Heathkit OL-1" you will find various references to it and the schematic, and several sources selling it. I wasn;t about to order the stuff myself.Education is what you're left with after you have forgotten what you have learned.
Comment
-
Originally posted by jasonguitar View Post. Hate to touch those cool old yellow spragues. Anyone restoring tweed or brown Fenders would go nuts for em!...
[added]
This may not be not be anything to do with your problem, but I picked up a scope a few years back, weird display that just didnt make any sense. I posted the problem to the Tekscopes group on Yahoo. A retired Tek engineer wrote back suggesting I clean the CRT socket - that fixed it. It's common for these guys when they get an old scope to wash them. Sounds weird, but at the Tek service centers, all scopes returned for service were first washed and then dried before any service or calibration."In theory, there is no difference between theory and practice. In practice there is."
- Yogi Berra
Comment
-
Originally posted by JoeM View Post...It's common for these guys when they get an old scope to wash them. Sounds weird, but at the Tek service centers, all scopes returned for service were first washed and then dried before any service or calibration.
Comment
-
Originally posted by Enzo View Post
Wait, you say you have all the documentation? Like including the schematic? Geez scan that and post it.
http://www.maintainband.com/AMP%20FILES/OL1_man.pdf
Comment
-
Here's a good article on the cleaning process used by TektronixAttached Files"In theory, there is no difference between theory and practice. In practice there is."
- Yogi Berra
Comment
-
Ah, terrific, schematics.
Look lower right corner, the B+ supply. Get out your meter and check all those B+ nodes, A, B, C, D. especially A. Are all those B+ DC voltages reasonable? Now flip the meter to AC volts and remeasure. We are measuring ripple now. A little ripple might be expected on A, but the rest should be real smooth and clean. Zero VAC, that is.Education is what you're left with after you have forgotten what you have learned.
Comment
-
Originally posted by Enzo View PostAh, terrific, schematics.
Look lower right corner, the B+ supply. Get out your meter and check all those B+ nodes, A, B, C, D. especially A. Are all those B+ DC voltages reasonable? Now flip the meter to AC volts and remeasure. We are measuring ripple now. A little ripple might be expected on A, but the rest should be real smooth and clean. Zero VAC, that is.
Comment
-
On page 24 of the manual, there's a voltage chart. I'd go through these first and check. Be careful, some high voltages in that thing."In theory, there is no difference between theory and practice. In practice there is."
- Yogi Berra
Comment
-
Originally posted by JoeM View PostOn page 24 of the manual, there's a voltage chart. I'd go through these first and check. Be careful, some high voltages in that thing.
Comment